"NB. It is very important to note that blood glucose concentrations measured using a whole blood
glucose meter calibrated for human blood may measure 30-40% lower in the low end of the range than
glucose concentrations measured using a serum chemistry analyser or a plasma-equivalent meter calibrated
for feline use. Therefore, if using a meter calibrated for feline use (eg. AlphaTRAK, Abbott
Laboratories, CA, USA), or a serum chemistry analyzer, add approximately 30 mg/dL (1.7 mmol/L)
to the target glucose concentrations (see Table 3B). For example, a target > 50 mg/dL (2.8
mmol/L) becomes > 80 mg/dL (4.4 mmo/L) when using a meter calibrated for feline use. Instead of
aiming for 50-100mg/dL (2.8-5.6 mmol/L) , aim for 80-130 mg/dL (4.4-7.2 mmol/L [round numbers
4.5-7.0 mmol/L). Meters calibrated for feline use may read higher or lower than the actual value, in
contrast to consistently lower readings for meters validated for human blood."
Larry and Kitties said:Regarding human meter reading intentionally low, a couple of years ago someone on the board contacted a few human meter manufacturers. They said there meters do not intentionally read low. Human meters are FDA approved devices and the manufacturers said it would be against FDA rules to have the meter intentionally read low.
Larry and Kitties said:What is your basis for saying the meters intentionally real low? I have never seen any facts that support that even though it is mentioned a lot. The meter manuals I have seen give a limit on the HCT for which the meter is accurate. The limit I have seen is outside the normal feline HCT levels. My One Touch Ultra 2 meter says it reads within HCT range of 30 t0 55%.
Larry and Kitties said:Regarding One Touch vs AlphTrac, how do you know that the One Touch reading is more accurate than the AlphaTrac and thus a higher low warning is required for the AlphaTrac. Maybe the AlphaTrac is more accurate and the error is with the One Touch.
That is correct, there is a slight difference and at least in the Lantus Group that has been pointed out (we have a FEW users with the AlphaTrak) those people are aware that there MINIMUM BG is higher than what the protocol (aka Tilly) recommends. They are also told to put clearly in thier SS and signature that they are using an AT meter so dosing advisor etc. can account for that fact.Actually, I'm NOT saying one is "more accurate" than the other. My point is that most people here use human meters and come to see 50s and 60s as "normal" (just look at the results that are seen in civvies). However, those same cats might test in the 70s or 80s on an AlphaTrak, meaning those values are "normal" for the AlphaTrak meter. It's a frame of reference issue, so to speak, and an AlphaTrak user needs to not take the commonly quoted "a hypo is below 40" or "a hypo is below 50" mantra too literally (IMHO).
It is very important to note that blood glucose concentrations measured using a whole blood
glucose meter calibrated for human blood may measure 30-40% lower in the low end of the range than
glucose concentrations measured using a serum chemistry analyser or a plasma-equivalent meter calibrated
for feline use. Therefore, if using a meter calibrated for feline use (eg. AlphaTRAK, Abbott
Laboratories, CA, USA), or a serum chemistry analyzer, add approximately 30 mg/dL (1.7 mmol/L)
to the target glucose concentrations (see Table 3B). For example, a target > 50 mg/dL (2.8
mmol/L) becomes > 80 mg/dL (4.4 mmo/L) when using a meter calibrated for feline use. Instead of
aiming for 50-100mg/dL (2.8-5.6 mmol/L) , aim for 80-130 mg/dL (4.4-7.2 mmol/L [round numbers
4.5-7.0 mmol/L). Meters calibrated for feline use may read higher or lower than the actual value, in
contrast to consistently lower readings for meters validated for human blood.