I've long had the FreeStyle Freedom Lite. After familiarizing myself with this board, I understood that FDMB users have had troubles with it under-reporting / being inaccurate.
So I did a side-by-side trial with a new Bayer Contour USB meter, which I borrowed from a family member (with human diabetes). I read it was recommended by a user or two of this site, but also noticed it was NOT very well rated in Consumer's Reports.
Copied below are 19 tests performed over 9 days between May 9th & May 18th. They were all drawn from the same blood sample at the same time. There were other readings in between where I only used my old FreeStyle meter because I couldn't get enough blood for two meters.
I was hoping to see that perhaps the FreeStyle would be on average either a certain number or certain percentage off (e.g. lower but within the 20% tolerance), but as you can see, it was not.
On average the difference was 62, or 42%.
You could almost infer that the longer I tested the two meters side-by-side, the larger the discrepancy in readings I was getting. But I think it was that the higher the reading, the larger the difference.
Is it possible that the FreeStyle was a bit low, but the Bayer was a bit high? One reading in particular, the Bayer was nearly double the FreeStyle... started me doubting the Bayer as well?
I've now returned the Bayer Contour USB and would like to buy another meter for more comparisons / likely replacing my FreeStyle.
I've always trusted Consumer's Reports, but on the other hand, it seems the cat experience of FDMB users doesn't necessarily align with human testing?
For example, the ReliOn and One Touch Ultra meters seem to be highly recommended here, but are only mid-pack in the CR ratings. Since I am in Canada, I don't think the ReliOn is available, so I would consider the OTU.
I haven't read much on the boards about the top-2 CR-recommended meters by Accu-Chek (Compact Plus or Aviva). Any luck with cats out there? All I've got to go on is my family member's recollection that they didn't like that brand much for human testing (error-prone and didn't last long a few years ago).
The one thing I do like about my current FreeStyle Freedom Lite meter (which is rated in the top-5 as per CR), is that the latest batch of strips I bought required a much smaller blood sample than the Bayer (despite the Bayer's 'sipping' action), so it was easier to get a reading. Other than that, all I can say is at least the FreeStyle seems to be on the same page with the other meter when the BG's get down towards hypo territory.
Thoughts?
FreeStyle Freedom Lite vs Bayer Contour USB
79 vs 79 (+0, 0%)
158 vs 184 (+25, 16%)
194 vs >600 (>405, >210%, assumed this was an error reading by the Bayer)
40 vs 38 (-2, -5%)
34 vs 36 (+2, 5%)
274 vs 396 (+122, 45%)
194 vs 277 (+83, 43%)
115 vs 142 (+27, 23%)
243 vs 362 (+119, 49%)
124 vs 155 (+31, 25%)
160 vs 247 (+86, 54%)
162 vs 203 (+41, 26%)
83 vs 106 (+23, 28%)
223 vs 437 (+214, 96%)
223 vs 369 (+146, 65%)
184 vs 265 (+81, 44%)
49 vs 50 (+2, 4%)
65 vs 74 (+9, 14%)
229 vs 378 (+149, 65%)
So I did a side-by-side trial with a new Bayer Contour USB meter, which I borrowed from a family member (with human diabetes). I read it was recommended by a user or two of this site, but also noticed it was NOT very well rated in Consumer's Reports.
Copied below are 19 tests performed over 9 days between May 9th & May 18th. They were all drawn from the same blood sample at the same time. There were other readings in between where I only used my old FreeStyle meter because I couldn't get enough blood for two meters.
I was hoping to see that perhaps the FreeStyle would be on average either a certain number or certain percentage off (e.g. lower but within the 20% tolerance), but as you can see, it was not.
On average the difference was 62, or 42%.
You could almost infer that the longer I tested the two meters side-by-side, the larger the discrepancy in readings I was getting. But I think it was that the higher the reading, the larger the difference.
Is it possible that the FreeStyle was a bit low, but the Bayer was a bit high? One reading in particular, the Bayer was nearly double the FreeStyle... started me doubting the Bayer as well?
I've now returned the Bayer Contour USB and would like to buy another meter for more comparisons / likely replacing my FreeStyle.
I've always trusted Consumer's Reports, but on the other hand, it seems the cat experience of FDMB users doesn't necessarily align with human testing?
For example, the ReliOn and One Touch Ultra meters seem to be highly recommended here, but are only mid-pack in the CR ratings. Since I am in Canada, I don't think the ReliOn is available, so I would consider the OTU.
I haven't read much on the boards about the top-2 CR-recommended meters by Accu-Chek (Compact Plus or Aviva). Any luck with cats out there? All I've got to go on is my family member's recollection that they didn't like that brand much for human testing (error-prone and didn't last long a few years ago).
The one thing I do like about my current FreeStyle Freedom Lite meter (which is rated in the top-5 as per CR), is that the latest batch of strips I bought required a much smaller blood sample than the Bayer (despite the Bayer's 'sipping' action), so it was easier to get a reading. Other than that, all I can say is at least the FreeStyle seems to be on the same page with the other meter when the BG's get down towards hypo territory.
Thoughts?
FreeStyle Freedom Lite vs Bayer Contour USB
79 vs 79 (+0, 0%)
158 vs 184 (+25, 16%)
194 vs >600 (>405, >210%, assumed this was an error reading by the Bayer)
40 vs 38 (-2, -5%)
34 vs 36 (+2, 5%)
274 vs 396 (+122, 45%)
194 vs 277 (+83, 43%)
115 vs 142 (+27, 23%)
243 vs 362 (+119, 49%)
124 vs 155 (+31, 25%)
160 vs 247 (+86, 54%)
162 vs 203 (+41, 26%)
83 vs 106 (+23, 28%)
223 vs 437 (+214, 96%)
223 vs 369 (+146, 65%)
184 vs 265 (+81, 44%)
49 vs 50 (+2, 4%)
65 vs 74 (+9, 14%)
229 vs 378 (+149, 65%)